APPENDIX A-2: Setup Variables For Serial I/O

SETUP VARIABLE
DAS_HOLD_OFF
TPC_ENABLE
RCELL_SET
1
IZS_SET
DYN_ZERO
DYN_SPAN
CONC_PRECISION
2
REM_CAL_DURATION
CLOCK_ADJ
SERVICE_CLEAR
TIME_SINCE_SVC
SVC_INTERVAL
1
T100/M100E
2
TAI protocol
06807F DCN7335
Table A-1:
NUMERIC
DEFAULT
UNITS
VALUE
Minutes
15
ON
ºC
50
Warnings:
45–55
ºC
50
Warnings:
45–55
OFF
OFF
3
Minutes
20
Sec./Day
0
OFF
Hours
0
Hours
0
Setup Variables
VALUE RANGE
0.5–20
Duration of DAS hold off period.
OFF, ON
ON enables temperature and pressure
compensation; OFF disables it.
30–70
Reaction cell temperature set point and
warning limits.
30–70
IZS temperature set point and warning
limits.
OFF, ON
ON enables contact closure dynamic
zero; OFF disables it.
OFF, ON
ON enables contact closure dynamic
span; OFF disables it.
AUTO,
Number of digits to display to the right
of the decimal point for concentrations
0,
on the display. Enclose value in double
1,
quotes (") when setting from the RS-
232 interface.
2,
3,
4
1–120
Duration of automatic calibration
initiated from TAI protocol.
-60–60
Time-of-day clock speed adjustment.
OFF
ON resets the service interval
timer.
ON
0–500000
Time since last service.
0–100000
Sets the interval between service
reminders.
DESCRIPTION
A-11