Data Acquisition System (DAS) and APICOM
7.2.11.

HOLDOFF FEATURE

Toggle to turn HOLDOFF
ON or OFF
164
The DAS HOLDOFF feature prevents data collection during calibrations and during the
DAS_HOLDOFF period enabled and specified in the VARS (refer to Section 5.8).
To enable or disable the HOLDOFF, follow the instruction shown in Figure 7-3, then
press:
Edit Data Channel Menu
SETUP X.X
0) CONC:
PREV NEXT
INS
SETUP X.X
NAME:CONC
<SET
SET>
EDIT
Press SET> until...
SETUP X.X
CAL HOLD OFF:ON
SET>
EDIT
PRINT
SETUP X.X
CAL HOLD OFF:ON
ON
Figure 7-12:
The DAS can be configured and operated remotely via the APICOM program.
Teledyne API - T100 UV Fluorescence SO
ATIMER,
2,
4032, R
DEL
EDIT
PRNT
EXIT
PRINT
EXIT
EXIT
ENTR
EXIT
DAS – Holdoff Feature
Analyzer
2
Exits to the main
Data Acquisition
menu
ENTR accepts the new
setting and returns to the
previous menu.
EXIT ignores the new setting
and returns to the previous
menu.