Table 6-7: Default Hessen Status Bit Assignments - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
6.6.2.5. HESSEN PROTOCOL GAS ID
6.6.2.6. SETTING HESSEN PROTOCOL STATUS FLAGS
06807F DCN7335
Analyzer
2
The T100 analyzer is a single gas instrument that measures SO
ID has already been set to 110. There is no need to change this setting.
Teledyne API's implementation of Hessen protocols includes a set of status bits that the
instrument includes in responses to inform the host computer of its condition. Each bit
can be assigned to one operational and warning message flag. The default settings for
these bit/flags are listed in Table 6-7:

Table 6-7: Default Hessen Status Bit Assignments

STATUS FLAG NAME
WARNING FLAGS
SAMPLE FLOW WARNING
PMT DET WARNING
UV LAMP WARNING
HVPS WARNING
DARK CAL WARNING
RCELL TEMP WARNING
IZS TEMP WARNING
PMT TEMP WARNING
INVALID CONC
OPERATIONAL FLAGS
In Manual Calibration Mode
In Zero Calibration Mode
In Span Calibration Mode
UNITS OF MEASURE FLAGS
UGM
MGM
PPB
PPM
SPARE/UNUSED BITS
UNASSIGNED FLAGS
Box Temp Warning
Sample Press Warning
System Reset
Rear Board Not Detected
Relay Board Warning
Communications Setup and Operation
DEFAULT BIT ASSIGNMENT
100. 8000
MP Calibration
Analog Cal Warning
Cannot Dyn Zero
Cannot Dyn Span
Instrument Off
. As such, its default gas
2
0001
0002
0002
0004
0008
0010
0020
0040
0080
0200
0400
0800
0000
2000
4000
6000
145