Figure 3-21: Pneumatic Layout With Izs Options - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
3.3.2.5. PNEUMATIC LAYOUT FOR INTERNAL ZERO/SPAN (IZS) GAS GENERATOR OPTION
EXHAUST GAS
OUTLET
SAMPLE GAS
INLET
ZERO/SPAN
VALVE
IZS
Permeation
Tube
SO
Source
2
ZERO AIR INLET
06807F DCN7335
Analyzer
2
The state of the zero/span valves can also be controlled by any of the following means:
manually from the analyzer's front panel by using the SIGNAL I/O controls
located within the DIAG Menu (refer to Section 5.9.1)
by activating the instrument's AutoCal feature (refer to Section 9.8)
remotely by using the external digital control inputs (refer to Section 8.1.2 and
Section 9.7.1)
remotely through the RS-232/485 serial I/O ports (refer to Appendix A-6 for the
appropriate commands)
Sources of zero and span gas must be capable of supplying at least 1.55 L/min.
(maximum 2.5L/min). The supply line should be vented outside of the analyzer's
enclosure. In order to prevent back-diffusion and pressure effects, these vent lines
should be between 2 and 10 meters in length.
Figure 3-21 shows the internal, pneumatic connections for the analyzer with the IZS
option installed.
KICKER EXHAUST
PUMP
SAMPLE/CAL
VALVE
COM
NO
NC
COM
NO
NC
Figure 3-21:
Pneumatic Layout with IZS Options
TO PUMP
HYDROCARBON
SCRUBBER
(KICKER)
EXHAUST TO OUTER LAYER
OF KICKER
CRITICAL
FLOW
ORIFICE
CRITICAL
FLOW
ORIFICE
Getting Started
Chassis
SAMPLE
CHAMBER
PMT
FLOW
SENSOR
SAMPLE
PRESSURE
SENSOR
FLOW / PRESSURE
SENSOR PCA
UV
LAMP
59