Remote Operation of the Analyzer
Table 8-1: Status Output Pin Assignments
CONNECTOR PIN
1
2
3
4
5
6
7-8
D
+
8.1.2.
CONTROL INPUTS
170
1
2
STATUS
System Ok
ON if no faults are present.
Conc Valid
ON if concentration measurement is valid, OFF when invalid.
High Range
ON if unit is in high range of any AUTO range mode.
Zero Cal
ON whenever the instrument is in ZERO calibration mode.
Span Cal
ON whenever the instrument is in SPAN calibration mode.
Diag Mode
ON whenever the instrument is in DIAGNOSTIC mode.
Unused
The emitters of the transistors on pins 1-8 are bussed together. For most
Emitter Bus
applications, this pin should be connected to the circuit ground of the
receiving device.
Dc Power
+ 5 VDC source, 30 mA maximum (combined rating with Control Inputs)
Digital Ground
The ground from the analyzer's internal, 5/±15 VDC power supply.
Control inputs allow the user to remotely initiate ZERO and SPAN calibration modes
provided through a 10-pin connector labeled CONTROL IN on the analyzer's rear
panel. These are opto-isolated, digital inputs that are activated when a 5 VDC signal
from the "U" pin is connected to the respective input pin.
Teledyne API - T100 UV Fluorescence SO
STATUS
3
4
5
6
7
Figure 8-1: Status Output Connector
CONDITION (ON=CONDUCTING)
+
8
D
06807F DCN7335
Analyzer
2