Single Range Mode (Sngl) - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Operating Instructions

4.8.3. Single Range Mode (SNGL)

SAMPLE*
RANGE = 500.000 PPM
< TST TST > CAL
PRIMARY SETUP MENU
SETUP C.3
CFG DAS
RNGE
PASS CLK MORE
SETUP C.3
RANGE CONTROL MENU
MODE
SET UNIT
SETUP C.3
RANGE MODE:
SNGL
DUAL AUTO
98
This is the default reporting range mode for the analyzer. In single range mode
both A1 and A2 are set to the same reporting range. This reporting range can be
any value between 50 ppm and 2 000 ppm.
While the two outputs always have the same reporting range, the span, signal
offset and scaling of their electronic signals may be configured for differently
(e.g., A1 = 0-10 V; A2 = 0-0.1 V). See Section 4.13.4 for instructions on
adjusting these parameters.
To select SNGL range mode and to set the upper limit of the range, press:
CO2 X.XXX
SETUP
EXIT
EXIT
SNGL
ENTR EXIT
Teledyne API Model T360/T360M Operation Manual
SETUP C.3
RANGE MODE:
SNGL DUAL AUTO
SETUP C.3
RANGE CONTROL MENU
MODE
SET
UNIT
SETUP C.3
RANGE:
500.0 Conc
0
0
5
0
0
SETUP C.3
RANGE CONTROL MENU
MODE SET UNIT
SNGL
ENTR
EXIT
EXIT
.0
ENTR
EXIT
EXIT
07272B DCN6552
EXIT x 2 returns
to the main
SAMPLE display