2.3.4.3 FUNCTIONAL CHECKS

After warm-up, verify that the software properly supports any hardware options that are
installed (Setup>Instrument menu), and that the instrument is functioning within
allowable operating parameters. Check the Dashboard page against the instrument's Final
Test and Validation Data sheet, which lists these values as they appeared before the
instrument left the factory. (If any functional parameters are not displayed, configure the
Dashboard through the Setup>Dashboard menu to add them; see Section 2.4.2).
These functions are also useful tools for diagnosing problems (information provided in
Section 5.7.2).

2.3.4.4 INITIAL CALIBRATION

Before operation begins, the analyzer requires initial calibrations for the location in
which it will perform any of the gas analyses. Also, any time an analyzer is moved or its
configuration changed, it must be calibrated. The method for performing a calibration
differs slightly depending on whether or not any of the available internal zero air or valve
options are installed. Follow the appropriate calibration instructions in Section 4.
083730400A DCN8130

Figure 2-24. Sample Dashboard Page