Calibration And Check Procedures With Valve Options Installed; Use Of Zero/Span Valve With Remote Contact Closure; Figure 4-2. Zero And Span Calibration Screens - Teledyne T200 User Manual

No/no2/nox analyzer with numaview software
Hide thumbs Also See for T200:
Table of Contents

Advertisement

C
C
P
V
O
ALIBRATION AND
HECK
ROCEDURES WITH
ALVE
PTIONS
I
NSTALLED

Figure 4-2. Zero and Span Calibration Screens

Follow the instructions in Section 4.2.1, except instead of the M-P menu, go to the
Calibration>Zero menu for Zero cal and to the Calibration>Span menu for NO Span cal.

USE OF ZERO/SPAN VALVE WITH REMOTE CONTACT CLOSURE

Contact closures for controlling calibration and calibration checks are located on the rear
panel CONTROL IN connector. Instructions for setup and use of these contacts are in
Section 2.3.1.6.
When the contacts are closed for at least 5 seconds, the instrument switches into zero, low
span or high span mode and the internal zero/span valves will be automatically switched
to the appropriate configuration.
98