4.13.8. Flow Calibration

Adjust these values
until the displayed
flow rate equals the
flow rate being
measured by the
independent flow
meter.
07272B DCN6552
The flow calibration allows the user to adjust the values of the sample flow rates
as they are displayed on the front panel and reported through COM ports to match
the actual flow rate measured at the sample inlet. This does not change the
hardware measurement of the flow sensors, only the software calculated values.
To carry out this adjustment, connect an external, sufficiently accurate flow meter
to the sample inlet (see Section 6.3.4 for more details). Once the flow meter is
attached and is measuring actual gas flow, press:
SAMPLE
RANGE = 500.000 PPM
< TST TST > CAL
SETUP X.X
PRIMARY SETUP MENU
CFG ACAL DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
SETUP X.X
ENTER DIAG PASS: 818
8
1
8
DIAG
SIGNAL I / O
NEXT
Repeat Pressing NEXT until . .
DIAG
FLOW CALIBRATION
PREV NEXT
DIAG FCAL
ACTUAL FLOW: 654 CC / M
0
6
5
4
CO2 =X.XXX
SETUP
MORE
EXIT
Exit at
any time
to return
to main
the
EXIT
SETUP
menu
ENTR
EXIT
ENTR EXIT
Adjust these values until
the displayed flow rate
equals the flow rate being
measured by the
independent flow meter.
.
Exit returns
ENTR EXIT
previous menu
ENTR accepts the
ENTR EXIT
new value and
returns to the
previous menu
EXIT ignores the
new value and
returns to the
previous menu