Analog Output Step Test - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Operating Instructions

4.6.2. ANALOG OUTPUT STEP TEST

SAMPLE
RANGE = 500.0 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
84
This test can be used to check the accuracy and proper operation of the analog outputs.
The test forces all four analog output channels to produce signals ranging from 0% to
100% of the full scale range in 20% increments. This test is useful to verify the operation
of the data logging/recording devices attached to the analyzer.
To begin the Anog Output Step Test press:
H2S
=XXX.X
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
Teledyne ML – T101 Operation Manual
DIAG
SIGNAL I / O
NEXT
DIAG
ANALOG OUTPUT
PREV
NEXT
DIAG AOUT
ANALOG OUTPUT
0%
DIAG AOUT
ANALOG OUTPUT
[0%]
Pressing the button under "0%" while performing the test will pause
the test at that level. Brackets will appear around the value; example:
[20%] Pressing the same button again will resume the test.
ENTR EXIT
ENTR
EXIT
Performs
analog output
step test.
0% - 100%
EXIT
Exit-Exit
returns to the
DIAG menu
EXIT