Operating Instructions
DIAG
Exit to return to
the main
PREV
sample display
DIAG AIO
< SET
Press SET> to select the analog output channel to
be configured. Then press EDIT to continue
DIAG AIO
< SET SET>
DIAG AIO
4.6.3.3. Manual Analog Output Calibration and Voltage Adjustment
Table 4-13. Voltage Tolerances for Analog Output Calibration
Full Scale
0.1 VDC
1 VDC
5 VDC
10 VDC
Outputs configured for 0.1V full scale should always be calibrated manually.
90
To select manual output calibration for a particular channel, activate the ANALOG I/O
CONFIGURATION MENU (see Section 4.6.3), then press:
ANALOG I / O CONFIGURATION
NEXT
ENTR
AOUTS CALIBRATED: NO
SET>
CAL
CONC_OUT_2:5V, CAL
EDIT
CONC_OUT_2 RANGE: 5V
SET>
EDIT
Now the analog output channels should either be automatically calibrated or they should
be set to manual calibration, which is described next.
For highest accuracy, the voltages of the analog outputs can be manually calibrated.
Calibration is done through the instrument software with a voltmeter connected across the
output terminals (Figure 4-4). Adjustments are made using the front panel buttons by
setting the zero-point first and then the span-point (Table 4-13).
The software allows this adjustment to be made in 100, 10 or 1 count increments.
Zero Tolerance
±0.0005V
±0.001V
±0.002V
±0.004V
Teledyne ML – T101 Operation Manual
DIAG AIO
EXIT
< SET
SET>
DIAG AIO
EXIT
< SET
SET>
DIAG AIO
ON
EXIT
Toggles the
auto cal mode
ON/ OFF for
this analog
output channel
only.
EXIT
Span Voltage
Span Tolerance
90 mV
±0.001V
900 mV
±0.001V
4500 mV
±0.003V
4500 mV
±0.006V
NOTE
CONC_OUT_2 REC OFS: 0 mV
EDIT
CONC_OUT_2 AUTO CAL: ON
EDIT
AOUT AUTO CAL: ON
ENTR EXIT
ENTR accepts the new setting
and returns to the previous
menu. EXIT ignores the new
setting and returns to the
previous menu.
EXIT