Figure 4-4. Setup For Calibrating Analog Outputs - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual
FROM ANALOG I/O CONFIGURATION MENU
DIAG
ANALOG I / O CONFIGURATION
PREV
NEXT
DIAG AIO
AOUTS CALIBRATED: NO
< SET
SET>
CAL
Press SET> to select the analog output channel to be configured:
DISPLAYED AS =
CONC_OUT_1 =
CONC_OUT_2 =
TEST OUTPUT =
DIAG AIO
CONC_OUT_1 :5V, NO CAL
< SET SET>
EDIT
Toggle to increase / decrease the analog output
Continue adjustments until the voltage measured
at the output of the analyzer and/or the input of
the recording device matches the value in the
upper right hand corner of the display to the
The concentration display will not change. Only
the voltage reading of your voltmeter will change.
See Table 3-1 for
pin assignments
of Analog Out
connector on the
rear panel
V OUT +
V OUT -
ANALYZER

Figure 4-4. Setup for Calibrating Analog Outputs

To make these adjustments, the AOUT auto-calibration feature must be turned off
(Section6.9).
Activate the ANALOG I/O CONFIGURATION MENU (see Section
4.6.3), then press:
ENTR
EXIT
EXIT
CHANNEL
A1
A2
A4
EXIT
by 100, 10 or 1 counts.
tolerance specified.
V
+DC
Gnd
V IN +
V IN -
Recording
Device
DIAG AIO
CONC_OUT_1 RANGE: 5V
SET>
EDIT
DIAG AIO
CONC_OUT_1 REC OFS: 0 mV
< SET
SET>
EDIT
DIAG AIO
CONC_OUT_1 AUTO CAL: OFF
< SET
SET>
EDIT
DIAG AIO
CONC_OUT_2 CALIBRATED: NO
< SET
CAL
DIAG AIO
CONC_OUT_1 VOLT–Z : 0 mV
U100 UP10 UP DOWN DN10 D100 ENTR EXIT
DIAG AIO
CONC_OUT_1 VOLT–S : 4500 mV
U100 UP10 UP DOWN DN10 D100 ENTR EXIT
DIAG AIO
CONC_OUT_1 CALIBRATED: YES
< SET
CAL
Operating Instructions
EXIT
EXIT
If AutoCal is ON, go to
Section 6.7.3
EXIT
EXIT
EXIT ignores the
new setting.
ENTR accepts the
new setting.
EXIT
91