Using The T101 With A Hessen Protocol Network; General Overview Of Hessen Protocol; Hessen Comm Port Configuration - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual

5.1.4. USING THE T101 WITH A HESSEN PROTOCOL NETWORK

5.1.4.1. General Overview of Hessen Protocol

5.1.4.2. Hessen COMM Port Configuration

Make sure that the communication parameters of the host computer are also properly
Also, the instrument software has 200 ms latency before it responds to commands
issued by the host computer. This latency should present no problems, but you should be
aware of it and not issue commands to the instrument too frequently.
The Hessen protocol is a Multidrop protocol, in which several remote instruments are
connected via a common communications channel to a host computer. The remote
instruments are regarded as slaves of the host computer. The remote instruments are
unaware that they are connected to a Multidrop bus and never initiate Hessen protocol
messages. They only respond to commands from the host computer and only when they
receive a command containing their own unique ID number.
The Hessen protocol is designed to accomplish two things: to obtain the status of remote
instruments, including the concentrations of all the gases measured; and to place remote
instruments into zero or span calibration or measure mode. TML's implementation
supports both of these principal features.
The Hessen protocol is not well defined; therefore while TML's application is completely
compatible with the protocol itself, it may be different from implementations by other
companies.
The following subsections describe the basics for setting up your instrument to operate
over a Hessen Protocol network. For more detailed information as well as a list of host
computer commands and examples of command and response message syntax, download
the Manual Addendum for Hessen Protocol from the Teledyne ML web site:
http://www.teledyne-ml.com/manuals/index.asp .
Hessen protocol requires the communication parameters of the T101's COMM ports to be
set differently than the standard configuration as shown in the table below.
Table 5-6. Hessen RS-232 Communication Parameters
Parameter
Data Bits
Stop Bits
Parity
Duplex
To change the rest of the COMM port parameters and modes, see Section 4.10.8.
To change the baud rate of the T101's COMM ports, see Section 4.10.9.
Standard
Hessen
8
7
1
2
None
Even
Full
Half
NOTE
set.
Remote Operation
155