Teledyne T101 Operation Manual page 112

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Operating Instructions
It is a good idea to check these settings the first time you power up your analyzer after it has been
physically connected to the LAN/Internet to make sure that the DHCP has successfully downloaded the
If the gateway IP, instrument IP and the subnet mask are all zeroes (e.g. "0.0.0.0"), the DCHP was not
You may have to manually configure the analyzer's Ethernet properties.
SAMPLE
RANGE = 500.0 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
PRIMARY SETUP MENU
SETUP X.X
CFG DAS RNGE PASS CLK
SECONDARY SETUP MENU
SETUP X.X
COMM
VARS DIAG
COMMUNICATIONS MENU
SETUP X.X
ID
INET
COM1
COM2
108
will automatically be assigned an IP address by the DHCP server (Section 4.10.6.2). This
configuration is useful for quickly getting an instrument up and running on a network.
However, for permanent Ethernet connections, a static IP address should be used. Section
4.7.6.2 below details how to configure the instrument with a static IP address.
appropriate information from you network server(s).
See your network administrator.
To view the LAN/Internet default configuration properties, press:
H2S
=XXX.X
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
EXIT
From this point on,
EXIT returns to
COMMUNICATIONS
MENU
Teledyne ML – T101 Operation Manual
NOTE
successful.
DHCP: ON
SETUP X.X
SET>
EDIT
INST IP: 0.0.0.0
SETUP X.X
<SET
SET>
GATEWAY IP: 0.0.0.0
SETUP X.X
<SET
SET>
SUBNET MASK: 0.0.0.0
SETUP X.X
<SET
SET>
TCP PORT: 3000
SETUP X.X
<SET
SET>
EDIT
HOSTNAME: T101
SETUP X.X
<SET
EDIT
EXIT
EXIT
EDITING
disabled
when
DHCP is
EXIT
ON
EXIT
EXIT
Do not alter
unless
directed to by
Teledyne
API's
Customer
Service
EXIT
personnel

Advertisement

Table of Contents
loading

Table of Contents