Calibration Procedures

Table 6-3. AutoCal Attribute Setup Parameters

Attribute Name
Timer Enabled
Starting Date
Starting Time
Delta Days
Delta Time
Duration
Calibrate
The programmed STARTING_TIME must be a minimum of 5 minutes later than the real
time clock (See Section 4.4.6 for setting real time clock).
Avoid setting two or more sequences at the same time of the day. Any new sequence
which is initiated whether from a timer, the COM ports, or the contact closure inputs will
If at any time an illegal entry is selected (Example: Delta Days > 367) the ENTR button
180
Turns on the Sequence timer
Sequence will operate on Starting Date
Sequence will operate at Starting Time
Number of days to skip between each sequence
Incremental delay on each Delta Day that the sequence starts.
Duration of the sequence in minutes
Enable to do dynamic zero/span calibration, disable to do a cal check only.
This must be set to OFF for units used in US EPA applications and with IZS
option installed.
override any sequence which is in progress.
will disappear from the display.