Table 6-4. Example Auto-Cal Sequence - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual

Table 6-4. Example Auto-Cal Sequence

MODE AND
ATTRIBUTE
SEQUENCE
MODE
TIMER ENABLE
STARTING DATE
STARTING TIME
DELTA DAYS
DELTA TIME
DURATION
CALIBRATE
The following example sets Sequence2 to carry out a zero-span calibration every other
day starting at 01:00 on September 4, 2002, lasting 15 minutes. This sequence will start
0.5 hours later each day.
VALUE
2
Define Sequence #2
ZERO-SPAN
Select Zero and Span Mode
ON
Enable the timer
Sept. 4, 2002
Start after Sept 4, 2002
01:00
First Span starts at 01:00
2
Do Sequence #2 every other day
00:30
Do Sequence #2 0.5 h later each day
15.0
Operate Span valve for 15 min
ON
The instrument will re-set the slope and offset values for the H
channel at the end of the AutoCal sequence
Calibration Procedures
COMMENT
S
2
181