4.8.3. ANALOG OUTPUT STEP TEST

SAMPLE
RANGE = 500.000 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
104
Analog Output is used as a step test to check the accuracy and proper operation of the
analog outputs. The test forces all four analog output channels to produce signals
ranging from 0% to 100% of the full scale range in 20% increments. This test is useful
to verify the operation of the data logging/recording devices attached to the analyzer.
Access the Analog Output Step Test from the DIAG Menu (refer to Figure 4-17), then
press:
SO2 =XXX.X
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
Figure 4-19:
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
DIAG
SIGNAL I / O
NEXT
DIAG
ANALOG OUTPUT
PREV
NEXT
DIAG AOUT
ANALOG OUTPUT
0%
DIAG AOUT
ANALOG OUTPUT
[0%]
Pressing the "0%" button while performing the test will
pause the test at that level. Brackets will appear around
the value: example: [20%] Pressing the same key again
will resume the test.
DIAG – Analog Output Menu
ENTR EXIT
ENTR
EXIT
Performs
analog output
step test.
0% - 100%
EXIT
Exit-Exit
returns to the
DIAG menu