Manually Configuring The Ethernet With Static Ip Addresses - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual

4.7.6.2. Manually Configuring the Ethernet with Static IP Addresses

1.
2.
3.
Table 4-17. LAN/Internet Default Configuration Properties
PROPERTY
DHCP
INSTRUMENT IP
ADDRESS
GATEWAY IP
ADDRESS
SUBNET MASK
1
TCP PORT
HOST NAME
1
Do not change the setting for this property unless instructed to by Teledyne ML's Technical
Support personnel.
Connect a cable from the analyzer's Ethernet port to a Local Area Network
(LAN) or Internet port.
Access Ethernet configuration through: SETUP>MORE>COMM>INET.
Follow the setup sequence as shown in the illustrations that follow, and
edit the Instrument and Gateway IP addresses and Subnet Mask to the
desired settings. (From the computer, enter the same information through
an application such as HyperTerminal).
DEFAULT STATE
This displays whether the DHCP is turned ON or
ON
OFF. Press EDIT and toggle ON for automatic
configuration after first consulting network
administrator.
This string of four packets of 1 to 3 numbers each
0.0.0.0
(e.g. 192.168.76.55.) is the address of the
analyzer itself.
Can only be edited when DHCP is set to OFF.
A string of numbers very similar to the Instrument
0.0.0.0
IP address (e.g. 192.168.76.1.) that is the address
of the computer used by your LAN to access the
Internet.
Can only be edited when DHCP is set to OFF.
Also a string of four packets of 1 to 3 numbers each
(e.g. 255.255.252.0) that identifies the LAN to
which the device is connected.
0.0.0.0
All addressable devices and computers on a LAN
must have the same subnet mask. Any
transmissions sent to devices with different subnets
are assumed to be outside of the LAN and are
routed through the gateway computer onto the
Internet.
This number defines the terminal control port by
3000
which the instrument is addressed by terminal
emulation software, such as Internet or Teledyne
ML's APICOM.
The name by which your analyzer will appear when
T101
addressed from other computers on the LAN or via
the Internet. To change, see Section 4.7.6.3.
Operating Instructions
DESCRIPTION
109