Teledyne T101 Operation Manual page 59

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual
The SPAN button now
appears during the
transition from zero to
span. You may see
both buttons.
If either the ZERO or
SPAN buttons fail to
appear see Section 11
for troubleshooting tips.
STEP THREE: Perform the zero/span calibration procedure:
SAMPLE
RANGE = 500.0 PPB
< TST
TST
>
CAL
SAMPLE
H2S STB=X.XXX PPB
< TST TST >
CAL
Allow zero gas to enter the sample port at the
rear of the instrument.
M-P CAL
H2S STB=X.XXX PPB
< TST TST >
CAL
M-P CAL
H2S STB=X.XXX PPB
< TST TST >
ZERO
M-P CAL
H2S STB=X.XXX PPB
< TST TST >
ENTR
Allow span gas to enter the sample port at the
rear of the instrument.
M-P CAL
H2S STB=X.XXX PPB
< TST TST >
SPAN
M-P CAL
RANGE = 500.0 PPB
< TST TST >
ENTR
SPAN CONC
M-P CAL
RANGE = 500.0 PPB
< TST TST > ENTR
The Model T101 analyzer is now ready for operation.
H2S
=XXX.X
SETUP
H2S =X.XXX
SETUP
ACTION:
SO2 =X.XXX
SETUP
SO2 =X.XXX
CONC
EXIT
SO2 =X.XXX
CONC
EXIT
ACTION:
H2S =X.XXX
CONC
EXIT
H2S =X.XXX
EXIT
H2S =X.XXX
CONC
EXIT
Getting Started
Set the Display to show the H2S
STB test function.
This function calculates the
stability of the H
S
2
measurement
Wait until H2S STB
falls below 0.5 ppb.
This may take several
minutes.
Press ENTR to edit the OFFSET &
SLOPE values for the H
S
2
measurements.
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
The value of
H2S STB may jump
significantly.
Wait until it falls back
below 0.5 ppb.
This may take several
minutes.
Press ENTR to change the offset &
slope values for the H
S
2
measurements.
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
EXIT returns to the main
SAMPLE display
55