Pressure Calibration - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Operating Instructions

4.6.7. PRESSURE CALIBRATION

SAMPLE
RANGE = 500.0 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
100
A sensor at the exit of the sample chamber continuously measures the pressure of the
sample gas. This data is used to compensate the final H
changes in atmospheric pressure when the instrument's TPC feature (see Section 10.7.3)
is turned on and is stored in the CPU's memory as the test function PRES (also viewable
via the front panel).
To cause the analyzer to measure and record a value for PRES, press:
H2S
=XXX.X
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
Teledyne ML – T101 Operation Manual
DIAG
NEXT
Repeat Pressing NEXT until . .
Exit at
any time
to return
to main
the
DIAG
SETUP
menu
PREV NEXT
DIAG PCAL
2
7
Adjust these values until the
displayed pressure equals the
pressure measured by the
independent pressure meter.
S concentration calculation for
2
SIGNAL I / O
ENTR
.
PRESSURE CALIBRATION
ENTR
ACTUAL PRES :27.20 IN-HG-A
.2
0
ENTR EXIT
ENTR accepts the
new value
EXIT ignores the new
value
EXIT
EXIT