Teledyne ML – T101 Operation Manual
The H
S span concentration
2
values automatically default
to 450.0 ppb.
To change this value to the
actual concentration of the
span gas, enter the number
by pressing the key under
each digit until the expected
value appears.
The span gas concentration
should always be 90% of the
selected reporting range
EXAMPLE
Reporting range = 800 ppb
Span gas conc.= 720 ppb
Step Two: Set the expected H
SAMPLE
RANGE = 500.0 PPB
< TST TST >
CAL
M-P CAL
RANGE = 500.000 PPB
< TST TST >
ZERO
M-P CAL
H2S SPAN CONC: 450.0 Conc
0
0
0
span gas
S
value:
2
H2S
=XXX.X
SETUP
H2S
=XXX.X
CONC
EXIT
4
5
0
.0
ENTR EXIT
Calibration Procedures
This sequence causes the
analyzer to prompt for the
expected H
S span
2
concentration.
EXIT ignores the new setting
and returns to the previous
display.
ENTR accepts the new setting
and returns to the
previous display..