9.2.1. ZERO OR LOW SAMPLE FLOW
9.2.2. HIGH FLOW
9.3. CALIBRATION PROBLEMS
9.3.1. NEGATIVE CONCENTRATIONS
If the pump is operating but the unit reports a 0 gas flow, do the following three steps:
Check for actual sample flow
•
Check pressure
•
Carry out a leak check
•
To check the actual sample flow, disconnect the sample tube from the sample inlet on
the rear panel of the instrument. Make sure that the unit is in basic SAMPLE mode.
Place a finger over the inlet and see if it gets sucked in by the vacuum or, more properly,
use a flow meter to measure the actual flow. If there is proper flow of around 550-650
cm³/min, contact Technical Support. If there is no flow or low flow, continue with the
next step.
Check that the sample pressure is at or around 26 in-Hg-A (about 1" below ambient
atmospheric pressure).
Flows that are significantly higher than the allowed operating range (typically ±10% of
the nominal flow) should not occur in the T101 unless a pressurized sample, zero or
span gas is supplied to the inlet ports. Be sure to vent excess pressure and flow just
before the analyzer inlet ports.
When supplying sample, zero or span gas at ambient pressure, a high flow would
indicate that one or more of the critical flow orifices are physically broken (very
unlikely case), allowing more than nominal flow, or were replaced with an orifice of
wrong specifications. If the flows are more than 15% higher than normal, we
recommend that the technician re-calibrate the flow electronically using the procedure in
Section 4.6.8, followed by a thorough and regular monitoring of these flows to see if the
new setting is retained properly.
This section presents some possible calibration problems and suggested solutions.
Negative concentration values can be caused for several things:
A slightly negative signal is normal when the analyzer is operating under
•
zero gas and the signal is drifting around the zero calibration point. This
is caused by the analyzer's zero noise and may cause reported
concentrations to be negative for a few seconds at a time down to -20
ppb, but should alternate with similarly high, positive values.
Mis-calibration is the most likely explanation for negative concentration
•
values. If the zero air contained some H
a worn-out zero air scrubber) and the analyzer was calibrated to that
concentration as "zero", the analyzer may report negative values when
measuring air that contains little or no H
the analyzer was zero-calibrated using ambient air or span gas.
Troubleshooting & Service
S gas (contaminated zero air or
2
S. The same problem occurs, if
2
221