Autocal Of Instruments In Independent Or Auto Reporting Range Modes - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Teledyne ML – T101 Operation Manual
6.9.1. AUTOCAL OF INSTRUMENTS IN INDEPENDENT OR AUTO
REPORTING RANGE MODES
It is recommended that, if dynamic calibration is enabled, the analyzer's test functions,
slope and offset values be checked frequently to assure high quality and accurate data
from the instrument.
If the analyzer is being operated in IND or AUTO Range mode, then the High and Low
ranges must be specified as part of the Auto Cal set up. This parameter appears at the end
of the programming sequences after the CALIBRATE: ON/OFF parameter is set. For
example:
Follow standard AutoCal programming process to this
point, then ...
SETUP X.X
CALIBRATE: ON
<SET
SET>
EDIT
SETUP X.X
RANGE TO CAL: LOW
<SET
EDIT
SETUP X.X
RANGE TO CAL: LOW
LOW
HIGH
SETUP X.X
RANGE TO CAL: HIGH
LOW
HIGH
SETUP X.X
RANGE TO CAL: HIGH
<SET
EDIT
SETUP X.X
SEQ 2) ZERO–SPAN, 2:00:30
PREV NEXT
MODE
SET
Sequence
EXIT
EXIT
ENTR
EXIT
ENTR
EXIT
EXIT
EXIT
Delta Time
MODE
Delta Days
Calibration Procedures
EXIT returns
to the SETUP
Menu
183