Setup - Acal: Auto Calibration; Setup - Das: Data Acquisition; Setup - Range: Analog Output Reporting Range Configuration; Available Analog Output Signals - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Operating Instructions
4.4.2. SETUP – ACAL: AUTO CALIBRATION
4.4.3. SETUP – DAS: DATA ACQUISITION
4.4.4. SETUP – RANGE: ANALOG OUTPUT REPORTING RANGE
CONFIGURATION

4.4.4.1. Available Analog Output Signals

LOW range when
DUAL mode is selected
66
Used to set up and operate the internal valve options if installed. Section 6 provides
details.
Used to set up the data acquisition system and record data.
The analyzer has three active analog output signals, accessible through a connector on the
rear panel.
H
S/SO
2
2
concentration outputs
A1
+
-
+

Figure 4-3. Analog Output Connectors Defined

All three outputs can be configured either at the factory or by the user for full scale
outputs of 0.1 VDC, 1VDC, 5VDC or 10VDC. Additionally A1 and A2 may be equipped
with optional 0-20 mADC current loop drivers and configured for any current output
within that range (e.g. 0-20, 2-20, 4-20, etc.). The user may also adjust the signal level
and scaling of the actual output voltage or current to match the input requirements of the
recorder or data logger (See Sections 4.6.3.3 and 4.6.3.5).
In its basic configuration, the A1 and A2 channels of the T101 output a signal that is
proportional to the H
S concentration of the sample gas. Several operating modes are
2
available which allow them to be slaved together (SNGL Mode, see Section 4.4.4.4 or
ANALOG OUT
A2
A3
A4
-
+
-
+
DUAL mode is selected
Teledyne ML – T101 Operation Manual
Not Used
Test Channel
-
HIGH range when