Calibration Procedures
Scroll to the H2S
STB test function.
Wait until H2S
STB falls below
0.5 ppb. This may
take several
minutes.
The value of H2S
STB may jump
significantly. Wait
until H2S STB falls
below 0.5 ppb. This
may take several
minutes.
176
STEP TWO: Perform the zero/span check.
SAMPLE
RANGE = 500.0 PPB
< TST TST >
CAL CALZ
SAMPLE
H2S STB=XXX.X PPB
< TST TST > CAL
ZERO CAL M
< TST TST >
ZERO
SAMPLE
H2S STB=XXX.X PPB
< TST TST > CAL
SPAN CAL M
H2S STB=XXX.X PPB
< TST TST >
ZERO SPAN
Teledyne ML – T101 Operation Manual
H2S =XXX.X
CALS
SETUP
H2S =XXX.X
CALZ
CALS
SETUP
H2S STB=XXX.X PPB
H2S =XXX.X
CONC
EXIT
H2S =XXX.X
CALZ
CALS
SETUP
H2S =XXX.X
CONC
EXIT
ACTION:
Record the
H
S readings
2
presented in the
upper right corner of
the display.
ACTION:
Record the
H
S readings
2
presented in the
upper right corner of
the display.
EXIT returns to the main
SAMPLE display