Operating Instructions

4.6.5. ELECTRICAL TEST

SAMPLE
RANGE = 500.0 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
98
The electrical test function creates a current, which substitutes the PMT signal, and feeds
it into the preamplifier board. This signal is generated by circuitry on the pre-amplifier
board itself and tests the filtering and amplification functions of that assembly along with
the A/D converter on the motherboard. It does not test the PMT itself. The electrical test
should produce a PMT signal of about 2000 ±1000 mV.
To activate the electrical test press:
H2S
=XXX.X
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
While the electrical test is
activated, PMT should equal:
2000 mV ± 1000 mV
Teledyne ML – T101 Operation Manual
DIAG
SIGNAL I / O
NEXT
Press NEXT until...
DIAG
ELECTRICAL TEST
PREV NEXT
DIAG ELEC
RANGE = 500.000 PPB
<TST TST>
Press TST until...
DIAG ELEC
PMT = 1732 MV
<TST TST>
ENTR
EXIT
ENTR
EXIT
O2=X.XXX
EXIT
SO2=X.XXX
EXIT