Operating Instructions
SAMPLE
RANGE = 500.0 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
PRIMARY SETUP MENU
SETUP X.X
CFG DAS RNGE PASS CLK
SECONDARY SETUP MENU
SETUP X.X
COMM
VARS
DIAG
SETUP X.X
0 ) MEASURE_MODE=H2S
NEXT
JUMP
76
In order to compensate for CPU clocks which run fast or slow, there is a variable to speed
up or slow down the clock by a fixed amount every day. To change this variable, press:
H2S
=XXX.X
SETUP
ENTR
EXIT
MORE
EXIT
EXIT
EDIT PRNT EXIT
Teledyne ML – T101 Operation Manual
SETUPX.X
1 ) DAS_HOLD_OFF=15.0 Minutes
PREV
NEXT
JUMP
Continue to press NEXT until ...
SETUP X.X
8) CLOCK_ADJ=0 Sec/Day
PREV
JUMP
SETUP X.X
CLOCK_ADJ:0 Sec/Day
+
0
0
Enter sign and number of seconds per
day the clock gains (-) or loses (+).
SETUP X.X
8) CLOCK_ADJ=0 Sec/Day
PREV NEXT JUMP
to the main SAMPLE display
EDIT PRNT EXIT
EDIT
PRNT EXIT
ENTR
EXIT
EDIT PRNT
EXIT
3x EXIT returns