Teledyne ML – T101 Operation Manual
Do not exceed 60 V between current loop outputs and instrument ground.
Similar to the voltage calibration, the software allows this current adjustment to be made
in 100, 10 or 1 count increments. Since the exact current increment per voltage count
varies from output to output and from instrument to instrument, you will need to measure
the change in the current with a current meter placed in series with the output circuit
(Figure 4-5).
See Table 3-2 for
pin assignments of
the Analog Out
connector on the
rear panel.
I OUT +
I OUT -
Analyzer
Figure 4-5. Setup for Calibrating Current Outputs
mA
IN
OUT
I IN +
I IN -
Recording
Device
NOTE
Operating Instructions
93