Setup - Pass: Password Protection; Table 4-6. Password Levels - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

Operating Instructions
4.4.5. SETUP – PASS: PASSWORD PROTECTION

Table 4-6. Password Levels

74
The menu system provides password protection of the calibration and setup functions to
prevent unauthorized adjustments. When the passwords have been enabled in the PASS
menu item, the system will prompt the user for a password anytime a password-protected
function (e.g., SETUP) is selected. This allows normal operation of the instrument, but
requires the password (101) to access to the menus under SETUP. When PASSWORD is
disabled (SETUP>OFF), any operator can enter the Primary Setup (SETUP) and
Secondary Setup (SETUP>MORE) menus. Whether PASSWORD is enabled or disabled,
a password (default 818) is required to enter the VARS or DIAG menus in the
SETUP>MORE menu.
PASSWORD
Null (000)
Operation
101
Configuration/Maintenance
818
Configuration/Maintenance Access to Secondary SETUP Submenus VARS
To enable or disable passwords, press:
SAMPLE
<TST
SETUP X.X
CFG DAS RNGE
SYSTEM
OFF
Toggle OFF to ON
to enable, and ON
to OFF to disable
SETUP X.X
password feature
ON
SETUP X.X
CFG DAS RNGE PASS CLK MORE
SAMPLE
<TST
SAMPLE
Toggle to change
password
LEVEL
All functions of the main menu (top level, or
Primary, menu)
Access to Primary and Secondary SETUP
Menus when PASSWORD is enabled
and DIAG whether PASSWORD is enabled or
disabled.
RANGE=xxx.x PPX
TST> CAL
PRIMARY SETUP MENU
PASS
CLK MORE
PASSWORD ENABLE: OFF
PASSWORD ENABLE: ON
PRIMARY SETUP MENU
RANGE=xxx.x PPX
TST> CAL
ENTER PASSWORD:0
0
0
0
Teledyne ML – T101 Operation Manual
MENU ACCESS ALLOWED
[GAS]= xxxx
SETUP
EXIT
ENTR EXIT
EXIT discards the new
ENTR EXIT
ENTR accepts the new
EXIT
[GAS]= xxxx
PRIMARY SETUP
SETUP
selecting SETUP
now prompts for
ENTR
EXIT
setting
setting
Press ENTR to
accept new
password, then
EXIT the
MENU so that
the password.