Table 7-1. Activity Matrix For Calibration Equipment & Supplies - Teledyne T101 Operation Manual

Uv fluorescence h2s analyzer
Table of Contents

Advertisement

EPA Protocol Calibration
Table 7-1. Activity Matrix for Calibration Equipment & Supplies
EQUIPMENT/SUPPLIES
Recorder
Sample Line And Manifold Constructed of PTFE or
Calibration Equipment
Working Standard SO
Cylinder Gas Or SO
Permeation Tube
Zero Air
Record Form
Audit Equipment
188
time. Therefore, it is necessary to dynamically check the calibration relationship on a
predetermined schedule. Zero and span checks must be used to document that the data
remains within control limits. These checks are also used in data reduction and validation.
Table 7-3 summarizes the initial quality assurance activities for calibrating equipment.
Table 7-2 is a matrix for the actual dynamic calibration procedure.
Calibrations should be carried out at the field monitoring site. The Analyzer should be in
operation for at least several hours (preferably overnight) before calibration so that it is
fully warmed up and its operation has stabilized. During the calibration, the T101 should
be in the CAL mode, and therefore sample the test atmosphere through all components
used during normal ambient sampling and through as much of the ambient air inlet
system as is practicable. If the Instrument will be used on more than one range (i.e. IND
or AUTO ranges), it should be calibrated separately on each applicable range. Calibration
documentation should be maintained with each analyzer and also in a central backup file.
ACCEPTANCE LIMITS
Compatible with output
signal of analyzer; min.
chart width of 150 mm (6
in) is recommended
glass
Meets guidelines of
reference 1 and Section
2.3.2 (Q.A. Handbook)
Traceable to NIST-SRM
2
meets limits in traceability
2
protocol for accuracy and
stability (Section 2.0.7,
Q.A. Handbook)
Clean dry ambient air, free
of contaminants that cause
detectable response with
the SO
analyzer.
2
Develop standard forms
Must not be the same as
used for calibration
Teledyne ML – T101 Operation Manual
FREQUENCY AND
ACTION IF REQUIREMENTS
METHOD OF
MEASUREMENT
Check upon receipt
Return equipment to
supplier
Check upon receipt
Return equipment to
supplier
See Section 2.3.9
Return equipment/ supplies
(Q.A. Handbook)
to supplier or take
corrective action
Analyzed against
Obtain new working
NIST-SRM; see
standard and check for
protocol in Section
traceability
2.0.7, Q.A.
Handbook
See Section 2.9.2
Obtain air from another
(Q.A. Handbook)
source or regenerate.
N/A
Revise forms as
appropriate
System must be
Locate problem and correct
checked out against
or return to supplier
known standards
ARE NOT MET