Operating Instructions
MODE
2
SETUP X.#
1
SPAN CAL A
1
SPAN CAL M
1
SPAN CAL R
1
ZERO CAL A
1
ZERO CAL M
1
ZERO CAL R
1
Only Appears on units with Z/S valve or IZS options
2
The revision of the analyzer firmware is displayed following the word SETUP, e.g., SETUP c.4

4.2. SAMPLE MODE

4.2.1. TEST FUNCTIONS

58
SETUP mode is being used to configure the analyzer. The gas measurement will
continue during this process.
Unit is performing SPAN calibration initiated automatically by the analyzer's
AUTOCAL feature
Unit is performing SPAN calibration initiated manually by the user.
Unit is performing SPAN calibration initiated remotely through the COM ports or
digital control inputs.
Unit is performing ZERO calibration procedure initiated automatically by the
AUTOCAL feature
Unit is performing ZERO calibration procedure initiated manually by the user.
Unit is performing ZERO calibration procedure initiated remotely through the COM
ports or digital control inputs.
Finally, the various CAL modes allow calibration of the analyzer. Calibration is described
in Section 6.
This is the analyzer's standard operating mode. In this mode, the instrument is analyzing
H
S
calculating concentrations.
2
and
A series of test functions is available at the front panel while the analyzer is in SAMPLE
mode. These parameters provide information about the present operating status of the
instrument and are useful during troubleshooting (Section 9.1.2). They can also be
recorded in one of the DAS channels (Section 4.8) for data analysis. To view the test
functions, press one of the <TST TST> buttons repeatedly in either direction.
Teledyne ML – T101 Operation Manual
DESCRIPTION