ALWAYS wear UV-Protective, Safety Glasses when working with the UV
DO NOT grasp the UV lamp by its cap when changing its position.

Figure 9-4. Shutter Assembly

26.
1. Set the analyzer display to show the signal I/O function,
UVLAMP_SIGNAL.
2. Slightly loosen the large brass thumbscrew located on the shutter
housing (refer to Figure 9-5) so that the lamp can be moved.
3. While watching the UVLAMP_SIGNAL reading, slowly rotate the lamp or
move it back and forth vertically until the UVLAMP_SIGNAL reading is
at its maximum.
Always grasp the main body of the lamp.
4. Compare the UVLAMP_SIGNAL reading to the information in Table 9-2
and follow the instructions there.