Teledyne API – Model T300/T300M CO Analyzer
Accuracy
Linearity
Temperature Coefficient
Rise and Fall Time
1
As defined by the USEPA
2
Defined as twice the zero noise level by the USEPA
2.2. EPA EQUIVALENCY DESIGNATION
Note
06864B DCN6314
<± (0.02% CO
+ 2% of reading)
2
<± 0.1% CO
2
<± 0.01% CO
/°C
2
<60 seconds to 95%
T300M: EPA equivalency does not apply to this model.
Teledyne API's Model T300, Carbon Monoxide Analyzer, is designated as Reference
Method Number RFCA-1093-093 as defined in 40 CFR Part 53, when operated under
the following conditions:
•
Range: Any range from 10 ppm to 50 ppm
•
Ambient temperature range of 10 to 40°C
•
Sample filter: Equipped with 5-micron PTFE filter element in the internal filter
assembly
•
Software settings:
Dilution factor
AutoCal
Dynamic Zero
Dynamic Span
Dual Range
Auto Range
Temp/Pres Compensation
Specifications and Approvals
1.0
ON or OFF
ON or OFF
OFF
ON or OFF
ON or OFF
ON