Manual Calibration Checks With Valve Options Installed - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Calibration Procedures
9.3.2. MANUAL CALIBRATION CHECKS WITH VALVE OPTIONS
INSTALLED
Front panel display
shows all of the
available gas
measurements
throughout this
procedure.
Wait until STABIL falls
below 0.2 PPM (for T300)
or 1.0 PPM (for T300M).
This may take several
minutes.
208
SAMPLE
RANGE=50.0PPM
< TST
TST >
CAL CALZ CALS
Toggle TST> button until ...
SAMPLE
STABIL= XXXX PPM
< TST TST >
CAL
CALZ CALS
Allow zero gas to enter the sample port
at the rear of the analyzer.
Record CO zero point readings
Allow span gas to enter the sample port
at the rear of the analyzer.
Record CO span point readings.
Teledyne API – Model T300/T300M CO Analyzer
CO=XX.XX
SETUP
CO=XXXX
SETUP
buttons will appear at various
It is not necessary to press
Set the Display to show
the STABIL test function.
This function calculates
the stability of the CO
measurement
Wait until STABIL falls
below 0.2 PPM (for T300)
or 1.0 PPM (for T300M).
This may take several
minutes.
The ZERO and/or SPAN
points of this process.
them.
06864B DCN6314