Adjusting The Sync/Demodulator, Circuit Gain - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Teledyne API – Model T300/T300M CO Analyzer

12.6.3.2. ADJUSTING THE SYNC/DEMODULATOR, CIRCUIT GAIN

Figure 12-19: Location of Sync/Demod Housing Mounting Screws
06864B DCN6314
To adjust the sync/demodulator circuit gain:
1. Make sure that the analyzer is turned on and warmed up.
2. Set the analyzer display to show the STABIL or CO STB test function.
3. Apply Zero Air to Sample Inlet of the analyzer.
4. Wait until the stability reading falls below 1.0 ppm.
5. Change the analyzer display to show the CO MEAS.
6. Remove the Sync/Demod Housing
Remove the two mounting screws.
Carefully lift the housing to reveal the sync/demod PCA.
7. Adjust potentiometer VR1 until CO MEAS reads 4500 mV ± 300 mV
VR1
Adjustment Made Here
Figure 12-20: Location of Sync/Demod Gain Potentiometer
Troubleshooting and Service
293

Advertisement

Table of Contents
loading

This manual is also suitable for:

T320T320uT300m