Teledyne API – Model T300/T300M CO Analyzer

5.9.3.10. AIN CALIBRATION

DIAG AIO
A/D CALIBRATION ERROR
Perform
Troubleshooting or call
Teledyne Customer
Service.
06864B DCN6314
This is the submenu to conduct a calibration of the T300/T300M Analyzer's analog
inputs.
This calibration should only be necessary after major repair such as a
replacement of CPU, motherboard or power supplies.
To perform an analog input calibration, select the ANALOG I/O CONFIGURATION
submenu (see Figure 5-3) then press:
AIO CONFIGURATION SUBMENU
DIAG
PREV NEXT
DIAG AIO
<SET
SET>
Continue pressing SET> until you reach the
DIAG AIO
AIN CALIBRATED:NO
<SET
DIAG AIO
Firmware automatically performs a zero point calibration
of the Motherboard's analog Inputs
DIAG AIO
Firmware automatically performs a span point calibration
of the Motherboard's analog Inputs
From the
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
CAL
CALIBRATING A/D ZERO
CALIBRATING A/D SPAN
DIAG AIO
EXIT
EXIT
EXIT
EXIT
AIN CALIBRATED: YES
Setup Menu
EXIT
135