Fault Diagnosis With Test Functions - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Troubleshooting and Service
Table 13-1: Warning Messages – Indicated Failures (cont.)
WARNING
MESSAGE
SAMPLE TEMP
Sample temperature is < 10
WARN
SOURCE WARNING
Either of these conditions will result
SYSTEM RESET
WHEEL TEMP
WARNING

12.1.2. FAULT DIAGNOSIS WITH TEST FUNCTIONS

Note
258
FAULT CONDITION
o
o
100
C.
Occurs when CO Ref is <1250
mVDC or >4950 mVDC.
in an invalid M/R ratio.
The computer has rebooted.
The filter wheel temperature is
controlled at 68 ± 2
°
C
Besides being useful as predictive diagnostic tools, the test functions viewable from the
front panel can be used to isolate and identify many operational problems when
combined with a thorough understanding of the analyzer's Theory of Operation (see
Section 13).
The acceptable ranges for these test functions are listed in the "Nominal Range" column
of the analyzer Final Test and Validation Data Sheet (T300, P/N 04307 and T300M,
P/N 04311) shipped with the instrument.
indicate a failure of one or more of the analyzer's subsystems. Functions whose values
are still within the acceptable range but have significantly changed from the
measurement recorded on the factory data sheet may also indicate a failure.
A worksheet has been provided in Appendix C to assist in recording the
value of these test functions. This worksheet also includes expected
values for the various test functions.
Teledyne API – Model T300/T300M CO Analyzer
Ambient temperature outside of specified range
Failed bench heater
C or >
Failed bench temperature sensor
Relay controlling the bench heater
Failed relay board
2
I
C bus
GFC Wheel stopped
Failed sync/demod board
If status LEDs on the sync/demod board ARE flashing the cause is
most likely a failed:
IR source
Relay board
2
I
C bus
IR photo-detector
This message occurs at power on. If you have not cycled the power
on your instrument:
Failed +5 VDC power,
Fatal error caused software to restart
Loose connector/wiring
Blocked cooling vents below GFC Assembly. Make sure that
adequate clear space beneath the analyzer.
Analyzer's top cover removed
Wheel heater
Wheel temperature sensor
Relay controlling the wheel heater
Entire relay board
2
I
C bus
Values outside these acceptable ranges
POSSIBLE CAUSES
06864B DCN6314