Calibration Procedures
9.7.1.2. SET O2 SPAN GAS CONCENTRATION
SAMPLE
RANGE=50.0 PPM
< TST TST >
CAL
SAMPLE
GAS TO CAL:CO
CO
CO2
O2
SAMPLE
GAS TO CAL:O2
CO
CO2
O2
224
Set the expected O
span gas concentration.
2
This should be equal to the percent concentration of the O
reporting range (default factory setting = 20.8%; the approximate O
air).
CO= XX.XX
SETUP
M-P CAL
ENTR EXIT
M-P CAL
ENTR
EXIT
The O
span concentration value automatically defaults to
2
20.8 %.
If this is not the the concentration of the span gas being
used, toggle these buttons to set the correct concentration
of the O
calibration gases.
2
Teledyne API – Model T300/T300M CO Analyzer
RANGE=50.0 PPM
<TST TST> ZERO SPAN
CONC
O2 SPAN CONC:20.95%
0
2
0
.9
5
span gas of the selected
2
content of ambient
2
CO= XX.XX
EXIT
EXIT ignores the new
ENTR EXIT
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the previous menu.