13.4.5.3. THERMISTOR INTERFACE

13.4.5.4. ANALOG OUTPUTS

13.4.5.5. INTERNAL DIGITAL I/O

13.4.5.6. EXTERNAL DIGITAL I/O

06864B DCN6314
This circuit provides excitation, termination and signal selection for several negative-
coefficient, thermistor temperature sensors located inside the analyzer. They are as
follows:
S
T
AMPLE
EMPERATURE
The source of this signal is a thermistor located inside the sample chamber of the Optical
Bench. It measures the temperature of the sample gas in the chamber. This data is used
to during the calculation of the CO concentration value.
B
T
ENCH
EMPERATURE
This thermistor is attached to the sample chamber housing. It reports the current
temperature of the chamber housing to the CPU as part of the bench heater control loop.
W
T
HEEL
EMPERATURE
This thermistor is attached to the heatsink on the GFC Wheel motor assembly. It reports
the current temperature of the wheel/motor assembly to the CPU as part of the Wheel
Heater control loop.
B
T
S
OX
EMPERATURE
ENSOR
A thermistor is attached to the motherboard.
temperature. This information is stored by the CPU and can be viewed by the user for
troubleshooting purposes via the front panel display (see Section 12.1.2).
The analyzer comes equipped with four analog outputs: A1, A2, A3 and A4. The type
of data and electronic performance of these outputs are configurable by the user (see
Section 5.4).
O
L
-
UTPUT
OOP
BACK
All four analog outputs are connected back to the A/D converter through a loop-back
circuit. This permits the voltage outputs to be calibrated by the CPU without need for
any additional tools or fixtures.
This channel is used to communicate digital status and control signals about the
operation of key components of the Optical Bench. The CPU sends signals to the
sync/demod board that initiate the ELECTRICAL TEST and DARK CALIBRATION
procedures.
This External Digital I/O performs two functions: status outputs and control inputs.
S
O
TATUS
UTPUTS
Logic-Level voltages are output through an optically isolated 8-pin connector located on
the rear panel of the analyzer. These outputs convey good/bad and on/off information
S
ENSOR
S
ENSOR
S
ENSOR
Theory of Operation
It measures the analyzer's internal