I 2 C Data Bus; Power Supply/ Circuit Breaker - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Theory of Operation
2
13.4.6. I
C DATA BUS

13.4.7. POWER SUPPLY/ CIRCUIT BREAKER

Should the AC power circuit breaker trip, investigate and correct the condition causing
320
about certain analyzer conditions. They can be used to interface with certain types of
programmable devices (See Section 3.3.1.4).
C
I
ONTROL
NPUTS
By applying +5VDC power supplied from an external source such as a PLC or Data
logger (See Section 3.3.1.6), Zero and Span calibrations can be initiated by contact
closures on the rear panel.
P
C
OWER UP
IRCUIT
This circuit monitors the +5V power supply during start-up and sets the analog outputs,
external digital I/O ports, and I
instrument software can establish control.
2
I
C is a two-wire, clocked, bi-directional, digital serial I/O bus that is used widely in
commercial and consumer electronic systems. A transceiver on the motherboard
converts data and control signals from the PC-104 bus to I
relay board, optional analog input board and valve driver board circuitry.
The analyzer operates on 100 VAC, 115 VAC or 230 VAC power at either 50Hz or
60Hz. Individual units are set up at the factory to accept any combination of these five
attributes. As illustrated in Figure 13-14, power enters the analyzer through a standard
IEC 320 power receptacle located on the rear panel of the instrument. From there it is
routed through the ON/OFF Switch located in the lower right corner of the Front Panel.
A 6.75 Amp circuit breaker is built into the ON/OFF Switch.
AC power is distributed directly to the sample gas pump. The bench and GFC Wheel
heaters as well as the GFC Wheel receive AC power via the relay board.
AC Line power is converted stepped down and converted to DC power by two DC
power supplies. One supplies +12 VDC, for valves and the IR source, while a second
supply provides +5 VDC and ±15 VDC for logic and analog circuitry. All DC voltages
are distributed via the relay board.
this situation before turning the analyzer back on.
Teledyne API – Model T300/T300M CO Analyzer
2
C circuitry to specific values until the CPU boots and the
CAUTION
G
S
H
ENERAL
AFETY
AZARD
2
C. The data is then fed to the
06864B DCN6314

Advertisement

Table of Contents
loading

This manual is also suitable for:

T320T320uT300m

Table of Contents