Teledyne API – Model T300/T300M CO Analyzer
Verify that the MODE
is set for SNGL.
If it is not, press
SINGL àENTR.
Verify that the RANGE is
set for 50.0
If it is not, toggle each
numeric button until the
proper range is set, then
press ENTR.
Verify that the UNIT
is set for PPM
If it is not, press
PPM àENTR.
06864B DCN6314
While these are the default setting for the T300/T300M Analyzer, it is recommended
that you verify them before proceeding with the calibration procedure, by pressing:
SAMPLE
<TST
TST> CAL
SETUP X.X
CFG DAS
SETUP X.X
MODE
SET
SETUP X.X
SNGL DUAL AUTO
SETUP X.X
MODE
SET
SETUP X.X
0
0
SETUP X.X
MODE
SET
SETUP X.X
PPB PPM UGM MGM
RANGE=50.0 PPM
CO= XX.XX
PRIMARY SETUP MENU
RNGE
PASS CLK MORE
RANGE CONTROL MENU
UNIT
DIL
RANGE MODE:SINGL
RANGE CONTROL MENU
UNIT
DIL
RANGE: 50.0 Conc
0
5
0
.0
RANGE CONTROL MENU
UNIT
DIL
CONC UNITS:PPM
Getting Started
SETUP
EXIT
EXIT
ENTR
EXIT
EXIT
ENTR
EXIT
Press EXIT as
needed to
return to
EXIT
SAMPLE
ENTR
EXIT
mode.
81