Teledyne API – Model T300/T300M CO Analyzer
SAMPLE
RANGE=50.0 PPM
<TST
TST> CAL
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
•
Pressing the button under "0%" pause
the test. Brackets will appear around the
value: EXAMPLE: [20%]
•
Pressing the same button again will
resume the test.
12.5.8.3. ANALOG OUTPUTS: CURRENT LOOP
06864B DCN6314
CO= XXXX
SETUP
MORE
EXIT
EXIT
Performs analog output step
test 0% to 100%
To verify that the analog outputs with the optional current mode output are working
properly, connect a 250 ohm resistor across the outputs and use a voltmeter to measure
the output as described in Section 5.9.3.6 and then perform an analog output step test as
described in Section 12.5.8.2.
For each step the output should be within 1% of the nominal value listed in the table
below.
Table 12-11:
Analog Output Test Function - Nominal Values Voltage Outputs
STEP
%
CURRENT
1
0
2 mA
2
20
5.6
3
40
9.2
4
60
12.8
5
80
16.4
6
100
20
SETUP X.X
ENTER PASSWORD
8
1
8
DIAG
SIGNAL I/O
PREV
NEXT
DIAG AOUT
ANALOG OUTPUT
20%
DIAG AOUT
ANALOG OUTPUT
[20%]
OUTPUT RANGE
2 -20
NOMINAL OUTPUT VALUES
V(250 OHMS)
CURRENT
0.5V
1.4
2.3
3.2
4.1
5
Troubleshooting and Service
ENTR
EXIT
ENTR
EXIT
EXIT
EXIT
4 -20
V(250 OHMS)
4
1
7.2
1.8
10.4
2.6
13.6
3.4
16.8
4.2
20
5
285