Teledyne API – Model T300/T300M CO Analyzer
Table 10-3:
Definition of Level 1 and Level 2 Zero and Span Checks
LEVEL 1 ZERO AND SPAN CALIBRATION
A Level 1 zero and span calibration is a simplified, two-
point analyzer calibration used when analyzer linearity
does not need to be checked or verified. (Sometimes
when no adjustments are made to the analyzer, the
Level 1 calibration may be called a zero/span check, in
which case it must not be confused with a Level 2
zero/span check.) Since most analyzers have a reliably
linear or near-linear output response with concentration,
they can be adequately calibrated with only two
concentration standards (two-point concentration).
Furthermore, one of the standards may be zero
concentration, which is relatively easily obtained and
need not be certified. Hence, only one certified
concentration standard is needed for the two-point (Level
1) zero and span calibration. Although lacking the
advantages of the multipoint calibration, the two-point
zero and span calibration--because of its simplicity--can
be (and should be) carried out much more frequently.
Also, two-point calibrations are easily automated.
Frequency checks or updating of the calibration
relationship with a two-point zero and span calibration
improves the quality of the monitoring data by helping to
keep the calibration relationship more closely matched to
any changes (drifts) in the analyzer response.
10.2. ZERO AND SPAN CHECKS
06864B DCN6314
1
(Q.A. Handbook
Vol II, Part1, Section 12.3 & 12.4)
A system of Level 1 and Level 2 zero span checks is recommended. These checks must
be conducted in accordance with the specific guidance given in Section 12 of the QA
Handbook Vol II Part 1
conducted every two weeks. Level 2 checks should be conducted in between the Level
1 checks at a frequency desired by the user. Span concentrations for both levels should
be between 70 and 90% of the measurement range.
Zero and span data are to be used to:
1. Provide data to allow analyzer adjustment for zero and span drift;
2. Provide a decision point on when to calibrate the analyzer;
3. Provide a decision point on invalidation of monitoring data.
Items 1 and 2 are described in detail in Subsection 9.1.3 of Section 2.0.9 (Q.A.
4
Handbook Vol II
). Item 3 is described in Subsection 9.1.4 of the same section.
Refer to the Troubleshooting and Repair (see Section 13) of this manual if the
instrument is not within the allowed variations.
LEVEL 2 ZERO AND SPAN CHECK
A Level 2 zero and span check is an "unofficial" check of
an analyzer's response. It may include dynamic checks
made with uncertified test concentrations, artificial
stimulation of the analyzer's detector, electronic or other
types of checks of a portion of the analyzer, etc.
Level 2 zero and span checks are not to be used as a
basis for analyzer zero or span adjustments, calibration
updates, or adjustment of ambient data. They are
intended as quick, convenient checks to be used
between zero and span calibrations to check for possible
analyzer malfunction or calibration drift. Whenever a
Level 2 zero or span check indicates a possible
calibration problem, a Level 1 zero and span (or
multipoint) calibration should be carried out before any
corrective action is taken.
If a Level 2 zero and span check is to be used in the
quality control program, a "reference response" for the
check should be obtained immediately following a zero
and span (or multipoint) calibration while the analyzer's
calibration is accurately known. Subsequent Level 2
check responses should then be compared to the most
recent reference response to determine if a change in
response has occurred. For automatic Level 2 zero and
span checks, the first scheduled check following the
calibration should be used for the reference response. It
should be kept in mind that any Level 2 check that
involves only part of the analyzer's system cannot
provide information about the portions of the system not
checked and therefore cannot be used as a verification
of the overall analyzer calibration.
1
. It is recommended that Level 1 zero and span checks be
EPA Calibration Protocol
235