Calibration Procedures

9.6.3. FLOW CALIBRATION

Toggle these buttons to
match the actual flow as
measured by the external
flow meter.
222
The flow calibration allows the user to adjust the values of the sample flow rates as they
are displayed on the front panel and reported through COMM ports to match the actual
flow rate measured at the sample inlet. This does not change the hardware measurement
of the flow sensors, only the software-calculated values.
To carry out this adjustment, connect an external, sufficiently accurate flow meter to the
sample inlet (see Section 11.3.4 for more details). Once the flow meter is attached and
is measuring actual gas flow, press:
SAMPLE
<TST
TST> CAL
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
SETUP X.X
ENTER PASSWORD:818
8
1
DIAG
PREV NEXT
Continue pressing NEXT until ...
DIAG
PREV NEXT
DIAG FCAL
1
0
See also Section
5.9.7.
Teledyne API – Model T300/T300M CO Analyzer
RANGE=50.0 PPM
CO= XX.XX
MORE
DIAG
8
ENTR
SIGNAL I/O
ENTR
FLOW CALIBRATION
ENTR
ACTUAL FLOW: 1000 CC/M
0
0
ENTR
SETUP
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT discards the new
setting.
ENTR accepts the
new setting.
06864B DCN6314