Pneumatic Layout For Zero Scrubber/ Ambient Span Option; Calibration Gases; Figure 3-28: Internal Pneumatic Flow Opt 50H - Zero Scrubber/Ambient Span; Table 3-11: Zero/Span Valve Operating States For Option 50H - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Getting Started

3.3.2.10. PNEUMATIC LAYOUT FOR ZERO SCRUBBER/ AMBIENT SPAN OPTION

Sample
SAMPLE
Gas
GAS INLET
In
Span
SPAN1
Gas
INLET
In
VENT SPAN
OUTLET
Zero
ZERO AIR
Gas
INLET
In
EXHAUST
GAS OUTLET
Figure 3-28:

3.3.2.11. CALIBRATION GASES

74
ZERO Air
Scrubber
Flow / Pressure
Sensor PCA
PUMP
Sample Gas
Flow Control
Internal Pneumatic Flow OPT 50H – Zero Scrubber/Ambient Span

Table 3-11: Zero/Span Valve Operating States for Option 50H

MODE
VALVE
SAMPLE
Sample/Cal
(Normal
Zero/Span
State)
Sample/Cal
ZERO CAL
Zero/Span
Sample/Cal
SPAN CAL
Zero/Span
Zero air and span gas are required for accurate calibration.
Z
A
ERO
IR
Zero air is a gas that is similar in chemical composition to the earth's atmosphere but
scrubbed of all components that might affect the analyzer's readings, in this case CO
and water vapor. If your analyzer is equipped with an Internal Zero Span (IZS) or an
external zero air scrubber option, it is capable of creating zero air.
Teledyne API – Model T300/T300M CO Analyzer
Sample / Cal
Valve
Particulate
Zero / Span
Valve
SAMPLE CHAMBER
SAMPLE
PRESSURE
SENSOR
FLOW
SENSOR
CONDITION
Open to SAMPLE inlet
Open to ZERO AIR scrubber
Open to ZERO/SPAN valve
Open to ZERO AIR scrubber
Open to ZERO/SPAN valve
Open to PRESSURE SPAN inlet
INSTRUMENT CHASSIS
Filter
GFC Motor Heat
Sync
GFC Wheel
Housing
06864B DCN6314