Maintenance Schedule & Procedures

11.3.4. PERFORMING A SAMPLE FLOW CHECK

Always use a separate calibrated flow meter capable of measuring flows in the 0 – 1000
DO NOT use the built in flow measurement viewable from the Front Panel of the
instrument. This measurement is only for detecting major flow interruptions such as

11.3.5. CLEANING THE OPTICAL BENCH

11.3.6. CLEANING EXTERIOR SURFACES OF THE T300/T300M

252
5. If the instrument has one of the zero and span valve options, the normally closed
ports on each valve should also be separately checked. Connect the leak checker
to the normally closed ports and check with soap bubble solution.
6. Once the leak has been located and repaired, the leak-down rate should be < 1 in-
Hg (0.4 psi) in 5 minutes after the pressure is shut off.
3
cm
/min range to measure the gas flow rate though the analyzer.
See Figure 3-4 for SAMPLE port location.
1. Attach the Flow Meter to the sample inlet port on the rear panel. Ensure that the
inlet to the Flow Meter is at atmospheric pressure.
2. Sample flow should be 800 cm
3. Once an accurate measurement has been recorded by the method described
above, adjust the analyzer's internal flow sensors (See Section 9.6.3).
Low flows indicate blockage somewhere in the pneumatic pathway, typically a plugged
sintered filter or critical flow orifice in one of the analyzer's flow control assemblies.
High flows indicate leaks downstream of the Flow Control Assembly.
The T300/T300M sensor assembly and optical bench are complex and delicate.
Disassembly and cleaning is not recommended. Please check with the factory before
disassembling the optical bench.
If necessary, the exterior surfaces of the T300/T300M can be cleaned with a clean damp
cloth. Do NOT submerge any part of the instrument and do NOT use any cleaning
solution.
Teledyne API – Model T300/T300M CO Analyzer
CAUTION
G
S
H
ENERAL
AFETY
AZARD
clogged or plugged gas lines.
/min ± 10%.
3
06864B DCN6314