Interferents For Co Measurements; Initial Calibration Procedure - Teledyne T300 Operation Manual

Carbon monoxide analyzer
Table of Contents

Advertisement

Getting Started

3.4.4.1. INTERFERENTS FOR CO MEASUREMENTS

3.4.4.2. INITIAL CALIBRATION PROCEDURE

80
It should be noted that the gas filter correlation method for detecting CO is subject to
interference from a number of other gases that absorb IR in a similar fashion to CO.
Most notable of these are water vapor, CO
The T300/T300M has been successfully tested for its ability to reject interference from
of these sources, however high concentrations of these gases can interfere with the
instrument's ability to make low-level CO measurements.
For a more detailed discussion of this topic, see Section 13.2.1.3.
The following procedure assumes that:
The instrument DOES NOT have any of the available calibration valve or gas inlet
options installed;
Cal gas will be supplied through the SAMPLE gas inlet on the back of the analyzer
(see Figure 3-4)
The pneumatic setup matches that described in Section 3.3.2.1.
V
T300/T300M R
ERIFYING THE
While it is possible to perform the following procedure with any range setting we
recommend that you perform this initial checkout using following reporting range
settings:
Unit of Measure: PPM
Analog Output Reporting Range: 50 ppm
Mode Setting: SNGL
Teledyne API – Model T300/T300M CO Analyzer
, N
O (nitrous oxide) and CH
2
2
R
S
EPORTING
ANGE
ETTINGS
(methane).
4
06864B DCN6314