Teledyne API – Model T300/T300M CO Analyzer
Only appears if either
the optional O2 or CO2
sensors are installed.
The CO span concentration value is
automatically default to
40.0 PPM.
If this is not the the concentration of
the span gas being used, toggle
these buttons to set the correct
concentration of the CO
calibration gas.
06864B DCN6314
S
CO S
G
C
ET
PAN
AS
ONCENTRATION
Set the expected CO pan gas concentration. This should be 80-90% of range of
concentration range for which the analyzer's analog output range is set.
SAMPLE
RANGE=50.0 PPM
<TST
TST>
CAL
SAMPLE
CO
O2
SAMPLE
LOW
HIGH
M-P CAL
RANGE=50.0 PPM
<TST TST> ZERO SPAN
M-P CAL
CO SPAN CONC:40.0 Conc
0
0
4
CO= XX.XX
GAS TO CAL:CO
ENTR
RANGE TO CAL:LOW
ENTR
CO= XX.XX
CONC
0
.0
0
ENTR EXIT
SETUP
EXIT
Only appears if the
analyzer is set for DUAL
EXIT
range (LOW/HIGH) mode
EXIT
EXIT ignores the new
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the
CONCENTRATION
MENU.