4.1.2. WARNING MESSAGES

Table 4-3:
List of Warning Messages
MESSAGE
ANALOG CAL WARNING
BENCH TEMP WARNING
BOX TEMP WARNING
CANNOT DYN SPAN
CANNOT DYN ZERO
CONC ALRM1 WARNING
CONC ALRM2 WARNING
CONFIG INITIALIZED
DATA INITIALIZED
O
CELL TEMP WARN
2
PHOTO TEMP WARNING
REAR BOARD NOT DET
RELAY BOARD WARN
SAMPLE FLOW WARN
SAMPLE PRESS WARN
SAMPLE TEMP WARN
SOURCE WARNING
1
SYSTEM RESET
WHEEL TEMP WARNING
1
Alarm warnings only present when 0ptional alarm package is activated.
2
Only enabled when the optional O
06864B DCN6314
The most common instrument failures will be reported as a warning on the analyzer's
front panel and through the COMM ports. Section 12.1.1 explains how to use these
messages to troubleshoot problems. Section 4.1.2 shows how to view and clear warning
messages.
The instrument's A/D circuitry or one of its analog outputs is not calibrated.
The temperature of the optical bench is outside the specified limits.
The temperature inside the chassis is outside the specified limits.
2
Remote span calibration failed while the dynamic span feature was set to turned on.
3
Remote zero calibration failed while the dynamic zero feature was set to turned on.
1
Concentration alarm 1 is enabled and the measured CO level is ≥ the set point.
1
Concentration alarm 2 is enabled and the measured CO level is ≥ the set point.
Configuration storage was reset to factory configuration or erased.
DAS data storage was erased.
2
O
sensor cell temperature outside of warning limits.
2
The temperature of the IR photo detector is outside the specified limits.
The CPU is unable to communicate with the motherboard.
The firmware is unable to communicate with the relay board.
The flow rate of the sample gas is outside the specified limits.
Sample gas pressure outside of
The temperature of the sample gas is outside the specified limits.
The IR source may be faulty.
The computer was rebooted.
The Gas Filter Correlation Wheel temperature is outside the specified limits.
Sensor is installed.
2
Overview of Operating Modes
MEANING
operational parameters.
93