Setup Menu

5.9.3.4. AUTOMATIC CALIBRATION OF THE ANALOG OUTPUTS

I
MPORTANT
CONC_OUT_3
is only active
on analyzers with
the optional O
2
sensor installed.
Analyzer
automatically
calibrates all
channels for which
AUTO-CAL is turned
ON.
If any of the channels have not
been calibrated or if at least one
channel has AUTO-CAL turned
OFF, this message will read NO.
I
MPORTANT
126
To calibrate the outputs as a group with the AOUTS CALIBRATION command, select
the ANALOG I/O CONFIGURATION submenu (see Figure 5-3) then press:
Before performing this procedure, make sure that the AUTO CAL for each
analog output is enabled. (See Section 5.9.3.3).
AIO CONFIGURATION SUBMENU
DIAG
ANALOG I/O CONFIGURATION
PREV NEXT
DIAG AIO
AOUTS CALIBRATED: NO
SET>
CAL
DIAG AIO
AUTO CALIBRATING CONC_OUT_1
DIAG AIO
DIAG AIO
DIAG AIO
DIAG AIO
SET> CAL
Manual calibration should be used for any analog output set for a 0.1V
output range or in cases where the outputs must be closely matched to
the characteristics of the recording device.
Teledyne API – Model T300/T300M CO Analyzer
I
R
MPACT ON
EADINGS OR
From the
ENTR
NOT AUTO CAL. CONC_OUT_2
AUTO CALIBRATING CONC_OUT_3
AUTO CALIBRATING TEST_OUTPUT
AOUTS CALIBRATED: NO
I
R
MPACT ON
EADINGS OR
D
ATA
EXIT
EXIT
This message
appears when
AUTO-CAL is
Turned OFF for
a channel.
EXIT
D
ATA
06864B DCN6314